- Posts: 10
- Joined: Wed Jan 23, 2013 9:05 pm
I'm reading a method validation report wherein the amount of compound (let's call it M) is calculated by using a response factor. Compound M is extracted into 1 mL of extraction solvent from 1 mL of 1:1 Acetonitrile:water solution. This is a GCMS method and they are calculating the response factor (RF) by using the conc. of compound M (mg/L)/peak height. They then use the formula: peak height x RF x (mL extraction solvent/mg compound M) to calculate the amount of compound M in ppm.
I'm having a hard time accepting this as a valid way to calculate the amount of compound M. I think it assumes a 100% extraction efficiency and also seems a little circular. Shouldn't there have been an internal standard used to calculate the RF instead? Or use an external standard of known concentration of compound M made up in the extraction solvent to truly calculate the amount of M in a sample as well as the extraction efficiency?
Thanks for your time!