This depends on the matrix. The typical strategy for a matrix that is easy to extract is to spike the analyte of interest into the matrix, mix well, perhaps evaporate the solvent used in spiking, and extract. If you can not obtain matrix that is analyte free, then measure spiked and unspiked matrix and subtract the response for the analyte in the unspiked sample from the response in the spiked sample.
I will often dothis kind of study by adding a recovery standard after the extraction and separation of the extraction solution from the matrix (works like an internal standard, but is not subject to the extraction process).
One can also extract for varying periods of time and look at the time trend. If one is extracting from a solid material where the analyte is in the solid and not on particle surfaces, one can try varying the particle size of the ground material to see if that changes the rate of extraction. One can also extract, remove the solvent and extract again repeatedlyl to estimate how complete the extraction is.
Thsi should give you a set of conditions that will work for all samples in the same matrix. If you must confirm extraction in all samples with variable matrix, the best may be to extract, remove the solvent and extract again. Analyze both extracts - but be sure that you have studied the extraction so that you know that the first extraction should be exhaustive for the given matrix, particle size, etc.