Page 1 of 1

Manual fraction collectors for HPLC (or DIY equivalents!)

Posted: Wed Jun 26, 2013 5:22 pm
by AaronAIT
Hi all,

Bit of a strange request, this one - would anyone know of a programmable fraction collector that doesn't necessarily have to be synced with chromatography software? I am asking on behalf of a colleague of mine, whose current method involves simply collecting the analyte from the HPLC waste line for ~30 seconds, and waiting 10 minutes to do so again. As you can imagine, this gets quite tedious.

They've tried controlling an old FPLC fraction collector without the FPLC connected, but had no such luck. Any ideas?


Cheers
Aaron.

Re: Manual fraction collectors for HPLC (or DIY equivalents!

Posted: Thu Jun 27, 2013 12:40 am
by tom jupille
If you're in the US, you can pick up old Pharmacia fraction collectors dirt-cheap from fleabay; e.g.:
http://www.ebay.com/itm/Pharmacia-Biote ... 53f9636d18

And those can collect on a time function, per this document:
http://web.mnstate.edu/provost/Pharmaci ... llInst.pdf

Re: Manual fraction collectors for HPLC (or DIY equivalents!

Posted: Thu Jun 27, 2013 2:45 pm
by ScottHorn
I recommend that you also take a look at the "Foxy 200" or any of it's re-branded equivalents (Rainin FC-1 comes to mind). You can find them used anywhere. They're pretty rugged machines and can be operated from the keypad manually, programmed to collect based on time or drop count, and if your detector has an analog voltage output as most do, you can run that into the fraction collector and have it collect all peaks or certain peaks without the need for any external PC controlling it. Most also come with a valve to divert to waste as well. I've bought several used ones and have yet to come across a lemon.


Scott Horn

Re: Manual fraction collectors for HPLC (or DIY equivalents!

Posted: Mon Jul 01, 2013 4:38 am
by Chris0000
Hello

yes, the Isco Foxy200 or the newer foxy R2 can do this, as they can be fully controlled by keypad.

Best regards Chris

Re: Manual fraction collectors for HPLC (or DIY equivalents!

Posted: Mon Jul 01, 2013 4:55 am
by AaronAIT
Thanks for the help lads - I'll get cracking on sourcing these asap!