Luna 5u c18(2) – 250*4.6mm column PROBLEMS
Posted: Wed Jul 27, 2005 8:00 pm
Hi ALL
Please HELP and sorry for such a long post
I have a problem with some chromatography using a Phenomenex Luna 5u c18(2) – 250*4.6mm column.
I am detecting for the following components:
Paracetamol - (10.1mins) - UV
Caffeine – (12.6) – UV
Phenylephrine HCL – (2.7) - FLD
Guaifenesin – (15.1) - FLD
Guaiacol - (16.0) – FLD
---------------------------------
METHOD is as follows:
The mobile phase: gradient system as follows:
Time A% B%
0mins 95 5
2mins 95 5
10mins 55 45
17mins 50 50
17.01mins 95 5
A - Water: formic acid 0.05%
B - Methanol:formic acid 0.05%
Run time: 20mins; Flow rate: 1.0ml/min; Injection: 3ul, Temp 35C
Detection: UV:275; FLD: ex 275nm em 300nm
Samples and standard made up in 0.1M hydrochloric acid (the samples have to be made up in this)
------------------------------------------------------------------------------
As you cam see from the following chromatograms.
http://www.davidwalkerhplc.zoomshare.com/1.shtml/HPLC
(1) In the first chromatogram the components are properly separated, with excellent Gaussian peaks
(2) The second chromatogram is the noise for the above standard: with slight retaining of the standards, but in general OK.
(3) The third Chromatogram is an appalling standard after only a few weeks column use.
(4) The fourth chromatogram is the noise that corresponds to the above standard
From these chromatograms it can be seen that over time the components are sticking to the column as the number of injections increase. As it can be seen from the noise of the last chromatogram.
So my question is: WHY?
Is it because the column is been destroyed by the low pH due to the samples/standards been made up in 0.1M HCL. If this is the case any suggestions of what column I could use to stop this?
OR - Is it the components are been ionized by the low pH of the 0.1M HCl.
If the components are been ionized, what buffer would you suggest I make up the samples/standards in?
Any suggestion would be fantastic help
Kind Regards
Dave
Please HELP and sorry for such a long post
I have a problem with some chromatography using a Phenomenex Luna 5u c18(2) – 250*4.6mm column.
I am detecting for the following components:
Paracetamol - (10.1mins) - UV
Caffeine – (12.6) – UV
Phenylephrine HCL – (2.7) - FLD
Guaifenesin – (15.1) - FLD
Guaiacol - (16.0) – FLD
---------------------------------
METHOD is as follows:
The mobile phase: gradient system as follows:
Time A% B%
0mins 95 5
2mins 95 5
10mins 55 45
17mins 50 50
17.01mins 95 5
A - Water: formic acid 0.05%
B - Methanol:formic acid 0.05%
Run time: 20mins; Flow rate: 1.0ml/min; Injection: 3ul, Temp 35C
Detection: UV:275; FLD: ex 275nm em 300nm
Samples and standard made up in 0.1M hydrochloric acid (the samples have to be made up in this)
------------------------------------------------------------------------------
As you cam see from the following chromatograms.
http://www.davidwalkerhplc.zoomshare.com/1.shtml/HPLC
(1) In the first chromatogram the components are properly separated, with excellent Gaussian peaks
(2) The second chromatogram is the noise for the above standard: with slight retaining of the standards, but in general OK.
(3) The third Chromatogram is an appalling standard after only a few weeks column use.
(4) The fourth chromatogram is the noise that corresponds to the above standard
From these chromatograms it can be seen that over time the components are sticking to the column as the number of injections increase. As it can be seen from the noise of the last chromatogram.
So my question is: WHY?
Is it because the column is been destroyed by the low pH due to the samples/standards been made up in 0.1M HCL. If this is the case any suggestions of what column I could use to stop this?
OR - Is it the components are been ionized by the low pH of the 0.1M HCl.
If the components are been ionized, what buffer would you suggest I make up the samples/standards in?
Any suggestion would be fantastic help
Kind Regards
Dave