Calibration of GC/MS Saturn Varian to analyse Pure Methanol

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

6 posts Page 1 of 1
Dear All,

I need to analyse pure Methanol (spécification is 99.5%) by GC/MS Saturn 2100T.

The problem I have is that when I inject the methanol, I have a very bad response (little pic of methanol)

The condition I use are :

Temperature of Injector : 200°C
Injector mode : Splitless

Colum temperature : 200°C during 5min then to 300°C @10°C per min

Volume injected = 0,5 uL

Could you help me please ?

Thank you very much
What is the lowest mass in your scan range ?

What is it that you need to know about the methanol ? its purity ?, or the identities and concentrations of impurities ?, or something else ?

Peter
Peter Apps
My lowest mass is 30.

I need to know the purity of the methanol and the identities adn concentrations of the impurities.

Thank you.
OK, you have strong ions at m/z 31 and 32.

Are you sure that the injection, inlet and column are working as they should ? How did you test them ?

The most likely impurity in methanol is water - you will not see that on the MS. How do you plan to determine the purity of them methanol ?, are you going to calibrate its peak against methanol standards of known purity ?

Just to be sure - do you have a solvent delay on the MS to protect the filaments ?

Peter
Peter Apps
I did not test the injection, column and inlet!
How I should process ?

I test the water by Kar Fisher.

I plan to determinate the purity of the methanol injecting 0.5 uL of the pure product, and I see other peaks, I will search the indentity in Saturn Library.

I have no solvent delay on the MS, because I inject the methanol pure (there is no solvent).
O.5 ul injected splitless will completely overload the MS; its electronics then automatically reduce the ion intensity in an attempt to avoid ion-ion interactions in the trap, and the end result is that no matter how much methanol you inject, or how pure it is, you will never see a large peak.

If you want to measure how much methanol you injected (which is what a purity determination requires) then you would need to do a split injection with a very high split ratio, and use an FID as the detector. I can save you the trouble of trying it by telling you that the variability in the peak areas will not allow you to distinguish between 99% and 100 % purity. There is also the problem that you cannot get 100.0 % pure methanol to use as a calibrant.

You could use the GC-MS to identify and quantify the impurities, but separating them from the methanol can be difficult.

Peter
Peter Apps
6 posts Page 1 of 1

Who is online

In total there are 30 users online :: 1 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry