MS oil contamination?

Posted: Fri Jun 21, 2013 2:41 pm
by BCBC
Hopefully the autotune picture will post. Power outage a few days ago. Pump stopped for a few minutes.
Believe this is rough pump oil contamination. Any thoughts? Thank you.


Image

Re: MS oil contamination?

Posted: Fri Jun 21, 2013 6:25 pm
by cleh
The tune post is very light so I can't read it on my device. Do you have a diff pump or turbo pump? It's definitely hydrocarbons, but if mass 446 is there, it's diff pump oil.

Re: MS oil contamination?

Posted: Wed Jun 26, 2013 6:45 pm
by Rick
Effectively, the post is very light.
If the peak are space by 14amu apart...Foreline pump oil
If 77, 94, 115, 141, 168, 170, 262, 354, 446......Diffusion pump oil

Re: MS oil contamination?

Posted: Thu Jun 27, 2013 11:23 am
by BCBC
Thanks. Appears to be a hydrocarbon signature to me too.

I'll check on the mass spacing.

I thought perhaps electronic noise but I believe that would be more of a grass look. Alot of lines, peaks close together, like grass.

Re: MS oil contamination?

Posted: Thu Jun 27, 2013 8:15 pm
by BCBC
Posting a new autotune. Hopefully it will post better than the previous one.

Have cleaned the MS manifold and analyzer. Replaced the ion source. Not sure what else may be contaminated. The ceramics?

Thank you.

Image

Re: MS oil contamination?

Posted: Thu Jun 27, 2013 8:34 pm
by Rick
You can suspect the column bleed too.....
For isolate this possibility, disconnect your column from interface and plug interface with ferrule with no hole...if you don't have it, you can install a short column piece (12inch) plug with a septum.

After that, pump the MS and reverify.

Re: MS oil contamination?

Posted: Fri Jun 28, 2013 6:26 am
by cleh
Thanks for the new tune post. Looks like rough pump oil to me. If it was column bleed you'd see mass 207. You may need new ceramics, but I'd clean or replace the vacuum hose going from the rough pump to the turbo or diff pump. It looks like you probably have a turbo pump and it is probably contaminated due to the back streaming so I would also clean it and replace the oil. Please let us know what you find.

Re: MS oil contamination?

Posted: Fri Jun 28, 2013 12:52 pm
by BCBC
This instrument has a diffusion pump. I did cap off the column from the GC and had the same pattern.

Thank you for the response. I'll replace the foreline hose and give it another look. Ceramics may need replacing if a new hose doesn't work.

Thank you.

Re: MS oil contamination?

Posted: Fri Jun 28, 2013 1:14 pm
by Rick
Ok..now, your ms is isolate.
You have clean the source..have you clean the Quad?...and have you clean the Mass spec chamber?...if yes, what solvant have you use?

You can try it:
- Verify the diffusion pump fan is operating
- Heat the source to 260c and the quad to 200c for 6hours (bake-out)
- Try with a new mechanical pump..sometimes, if you have a small leak, valve have leak and produce a mist oil
- Change ou clean the electron multiplier (isopropanol for cleaning and dry it before reinstall)

Re: MS oil contamination?

Posted: Sat Jun 29, 2013 2:34 pm
by BCBC
Thank you. I have rinsed the Quad and associated metal parts with Methylene Chloride. Replaced the Ion Source. Replace the Multiplier. Wiped the Analyzer housing with Methylene Chloride. The autotune looks exactly as the one previously posted.

Perhaps the ceramics and the items you suggested, fan, heating etc.

Re: MS oil contamination?

Posted: Sat Jun 29, 2013 4:09 pm
by cleh
I would clean the diff pump and change the oil. You can check the fan and clean or replace if needed. Clean the transfer line tip, also.

Re: MS oil contamination?

Posted: Sat Jun 29, 2013 10:12 pm
by Don_Hilton
Back some years ago, we had a rough pump fail in a HP5972. We lifted the analyzer section off the top of the mass spec, removed the source and detector, and poured methylene chloride across the whole unit -- rinsing quads, mounting hardware, the plate holding everything, and all. We rinsed out the chamber as well. I think we discarded the detector and replaced it. And, we did clean the source. The diffusion pump oil had to be changed out because it was contaminated.

On pumping down we had a huge background of methylene chloride for about a week - but after that, we had a clean mass spec that we put back into service.

Re: MS oil contamination?

Posted: Sun Jun 30, 2013 4:39 am
by cleh
Agree with Don.

Re: MS oil contamination?

Posted: Mon Jul 01, 2013 11:15 am
by BCBC
Thank you for all the suggestions. I'll go through and flush all that I can with Methylene Chloride.

Also will change the diffusion pump oil. Makes sense that it could be contaminated; wouldn't be surprised.

Cleaning probably won't take place until next week.

Thank you again.

Re: MS oil contamination?

Posted: Mon Jul 01, 2013 12:04 pm
by LCbob
I would also check your rough pump to make sure it's anti siphon ( prevents back streaming ) is working. You don't want to do all this work and the same happens again!