Seven residual solvents, Peaks missing!!!!
Posted: Wed Jun 19, 2013 4:52 pm
Hi all,
I am having some issues trying to write an analytical method to combine seven different residual solvents together in a diluent to establish seven peaks to be visible by headspace GC. The column i'm using is Durabond DB-WAX 30 m x 0.530 mm, 1 ufilm.
The seven residual solvents and their concentrations are: 250 ppm Acetone, 100 ppm Methanol, 30 ppm Dichloromethane, 20 ppm Toluene, 18 ppm Chlorobenzene, 16 ppm Triethylamine, and 250 ppm Acetic Acid.
I prepared a standard in purified water with them all however, only five peaks showed up (acetone, methanol, dichloromethane, toluene, and chlorobenzene). Triethylamine and Acetic Acid did not have peaks present, i think this is due to TEA and Acetic Acid reacting in water.
Next, i prepared TEA and Acetic Acid individually in purified water to see if the peaks would show up. TEA concentration was 14.935 ppm and Acetic Acid concentration was 249.25 ppm. Separately, i got a shift in retention time for TEA at 3.098 (before when i injected TEA in 2 mL purified water i got a retention time of 2.339) and a peak area of 8.5 pA*s (which to me does not sound right), Acetic acid also had a small peak at 13.816 retention time and a peak area of 8.7 pA*s. When combining the two with purified water, neither peak showed up.
I have concluded that these two solvents cannot be quantified in purified water, probably because TEA-acetate is produced??
Then, I prepared individual vials of 1 drop TEA + 2 mL DMF, 1 drop Acetic Acid + 2 mL DMF, and i prepared one vial containing 1 drop of TEA and 1 drop Acetic Acid + 2 mL DMF. Individually, TEA had a retention time at 2.399 and Acetic Acid at 13.714. Combined, both peaks showed up too at roughly the same retention time (2.401, 13.749 respectively).
Thinking I'm all good, i prepared a standard solution with all residual solvents in DMF. I prepared the standard with the same concentrations as above...yet again when i came in the morning, i had missing peaks, this time I was missing three/four peaks ( i say that cause when i zoomed in at the retention time where toluene should be, i discovered a tiny tiny peak and Toluene showed up for the 120% linearity with a peak area of 14 pA*s at a concentration of 24 ppm). The peaks present were Acetone, Methanol, and chlorobenzene (DMF was detected at retention time 12.913). The areas of these peaks were definitely not as high as they were in purified water but i'm not too sure how relevant that is. TEA, Dichloromethane, and Acetic Acid were missing.
Next i tried the same preparation except in DMSO. Individually, 1 drop Acetic Acid + 2 mL DMSO, 1 drop TEA + 2 mL DMSO, 1 drop chlorobenzene + 2 mL DMSO, 1 drop TEA + 1 drop Acetic Acid + 2 mL DMSO, and 1 drop TEA + 1 drop Acetic Acid + 1 drop chlorobenzene + 2 mL DMSO. All peaks showed up, even for the two vials that had them combined. HOWEVER!! when all seven solvents were combined together in DMSO with the concentrations above, there was an overlap of what i think would be TEA and Acetone (2.738, 2.876 respectively). Dichloromethane, and Acetic Acid did not show up!
I do not know how to write a method for all seven of these residual solvents if peaks are missing and its constantly the same peaks that are missing. Please help! Can i just not combine all seven of these solvents. Should i separate two different methods and have some residual solvents in one method and others in another method. I know i can validate Acetone, Methanol, Dichloromethane, and Toluene together (maybe even Cholorbenzene), but what do I do with TEA and Acetic Acid.
Please let me know if you have any questions or have any answers for me.
Thanks!
Thanks!