Gilson HPLC with agilent detector
Posted: Wed Jun 19, 2013 10:46 am
by cottrel
Dear all,
do any of you know if this is possible? we are trying to use a fluorescence detector from on a Gilson HPLC system, is there a way to control the fluorescence detector and the hplc via some software?
Thanks for your help,
Cottrel
Re: Gilson HPLC with agilent detector
Posted: Fri Jun 28, 2013 9:03 am
by elsico
You can control Jasco 920 (Gilson 122) detector from Unipoint software. But you need Gilson 606 GSIOC To RS232 Converter.
Posted: Thu Jul 04, 2013 10:47 pm
by cottrel
Thanks for your reply.
Actually it turned out that it's not an Agilent but a Jasco fp-2020 fluorescence detector. Apologies for my dumb question but how do I connect it to the gilson HPLC interface module on a very basic level? I am using trilution as software.
Cottrel
Posted: Fri Jul 05, 2013 2:17 am
by tom jupille
I hate to say this, but your best bet would be to contact Gilson. Trilution is really aimed at controlling fraction collection; I took a look at the user manual (800 pages!), and it pretty much assumes you're using Gilson hardware.
Posted: Fri Jul 05, 2013 9:23 am
by cottrel
Thanks Tom,
I have contacted customer support...waiting for a reply...
Posted: Wed Jul 17, 2013 8:05 am
by cottrel
Just to summarise how it went.
I contacted Gilson directly. It was easy in the end. I just connected the "rec out" at the back of the detector to the controller of the Gilson making sure to keep the polarities. Once done that, I made sure to enable the same channel I connected the detector to on the software (trilution in this case).Simple as that.
Jasco customer service also provided the manual for the detector, I could do a few more things with the other connectors but for now just being able to see a signal is sufficient.
Regards