Hi Terry,
There isn't a "kind" of HPLC, per se. If the instrument has a "flow through needle," where the needle and the sample loop are in the eluent flow path during the chromatographic run, then the only limit to the injection volume is the loop volume itself.
If the instrument has a "fixed loop," then the loop is segregated from the eluent flow path during the chromatographic run. The sample "slug" is pushed, or pulled, into the eluent flow path...eluent is not sweeping the volume of the sample loop during the run.
In general, high-pressure mixing systems employ "fixed loop" design, though there are exceptions...and low-pressure mixing designs generally employ "flow through needle) design, though again, there are exceptions.
In the Waters world...the Acquity-H is a flow through needle, and both flow through needle AND fixed loop are employed in the Acquity-I...you can see both at Waters Web Site (fixed loop first, then flow through needle). Both Acquity-I instruments are binary high-pressure mixing, though one has a fixed loop and the second a flow through needle.
http://www.waters.com/waters/library.ht ... =134619338
http://www.waters.com/waters/library.ht ... =134619288
The difference...there will be less gradient delay volume in the fixed loop design...fixed loop has a bit of an advantage on that count. Flow through needle "ought" to have an advantage in sample carryover, as the loop is multiply flushed.
Does this help a bit? You can find an explanation of the fixed sample loop injector at the IDEX web site...Rheodyne has been making this design for many years in their manual injection valves...as well as their automated ones commonly employed in Dionex ICs and Thermo Separation Products LCs. I'll send a link to this thread later on today if I can--see below after an edit...
http://www.idex-hs.com/support/rheodyne ... note_5.pdf