Solvent front not always the same

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
We run a HPLC Dissolution method that can show large variations in the solvent front we get on the chromatography. We have narrowed it down to the Sodium Hydroxide but it will differ even if the same brand of the reagent is used. Has anyone had this problem before? From a regulatory point of view will it be a problem if the chromatograms have different solvent fronts on the chromatography between batches but the peaks of interest are OK?

Thanks in advance
"t0 noise" from unretained junk is common and as you have experienced, can be quite variable.
From a regulatory point of view will it be a problem if the chromatograms have different solvent fronts on the chromatography between batches but the peaks of interest are OK?
Does your method include anything about t0 in the system suitability criteria? So long as you meet system suitability, your system is suitable.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thanks Tom, is it not strenge that every thing is the same but the chromatogram looks different?
I wouldn't be concerned over things that are of no concern.

That sounds like the master chromatographist Yogi Berra said that, but there's enough real world stuff to worry about.

Code: Select all

Thanks Tom, is it not strange that every thing is the same but the chromatogram looks different?
Not really. You start off with a chromatographic system at steady-state (i.e., close to equilibrium) and then you disturb the equilibrium by making an injection. Until the system returns to equilibrium, it's in a chaotic state, with everything that implies about (non)repeatability and (non)predictability. That's why the FDA Guidance suggests that quantitation should be done on peaks with k'> 2; that ensures that they are outside that chaotic region.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
5 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry