TotalChrom error during run setup

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
I am working with a re-installation of TotalChrom workstation v6.3. I am preparing for validation, but I cannot setup a run. When I try to setup from a sequence, I get the following error: -995 Couldn't open seqfile. Is this a file permissions issue? I have tried a new sequence and a previously-used. Same error every time.
Hi, Kate,

It indicates that some set up file are missing, so needs to contact with service statation, they only can solve the problem.

Take care
Bye
To close this topic, I will post the resolution. This was a file permissions issue. After consulting with Perkin Elmer, I asked our IT to re-check the permissions matrix established during the install. There were errors. Fixing the file permissions issues solved the problem :-)
3 posts Page 1 of 1

Who is online

In total there are 285 users online :: 1 registered, 0 hidden and 284 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 284 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry