We are working in a 7890A GC connected to 5975C MS and we are analyzing PCB's with a DB-5 column in ground water and VOC's with an HP-VOC column. Before the sample collection we have standardized both of the methods and made calibration curves with calibration standards and we were waiting to collect the samples and when the samples were collected we analyzed first VOC's because they are more volatile than PCB's.
When we installed the DB-5 column an electricity brake down happened for hole night and both of the filaments burned. We changed one of them (the second ordered, didn't come yet) and when we tried to run a standard, the chromatogram was far from what we expected. With the same oven program, the same compounds and concentration, same column, everything same, the retention times has changed, some compounds are co-eluting and some others disappeared, and the size of the peeks is getting smaller as the session in moving.
Has anyone any idea about what might be the cause? is the filament change? is the fact that we haven't installed both of filaments? maybe something else happened in the instrument?
I am posting you photos of TIC before and after the filament change to understand better...
before

and after

Hope someone will have the answer...
Thanks in advance..!!!