hplc calibration
Posted: Sun Jun 16, 2013 11:29 am
by AHSJ1269
in HPLC calibration wave length accuracy
how i determine wave length accuracy of detector at lower uv wave lenghts i.e. 210 nm , 205 nm
Re: hplc calibration
Posted: Sun Jun 16, 2013 3:05 pm
by tom jupille
You don't. Almost any signal you see in that region is "end absorbance" which continues to increase as you go to shorter wavelengths (i.e., hard to find anything that has a true absorbance maximum in that region).
Re: hplc calibration
Posted: Tue Jun 18, 2013 10:24 am
by lmh
I sort-of feel that wavelength calibration is one of the less useful calibrations we do anyway, at least in quantitative work.
Most things have fairly broad maxima, so if you're out by a few nm, it won't make any difference. And in any case, if the standards and samples are both measured at the same wavelength, the actual exact wavelength used probably doesn't matter. If you're comparing wavelength maxima to expected values as part of a strategy for identifying things, that of course is a whole different situation.