3u should save some solvent and give better resolution for its length, so you're certainly thinking along sensible lines. However, there are some other things you could try too:
(1) If you're worried about solvent, have you considered reducing column diameter? A 2mm column has 1/5th of the cross-sectional area of a 4.6mm column, so you can reduce the flow-rate from, say, 1mL/min to 200uL/min, thereby saving huge amounts of solvent, but with minimal change to your method. The only considerations are that the dead volume of your system will delay things slightly more (very old HPLC systems may struggle, but anything vaguely modern will be fine), and that you may need to consider reducing injection volume if you're injecting in a solvent that is stronger than the running solvent, or if the sample is nearly overloading. In general, a narrow-bore column may allow you to increase sensitivity as many detectors (e.g. PDA/UV) measure concentration, not amount.
(2) Have you considered fused-core particles? My (limited) experience is that whatever your Waters rep tells you to the contrary, and however much theory he has to back it up, they actually do work. They give many of the advantages of very small particles, but with the back-pressure associated with larger particles. I don't want to say bad things about UPLC and tiny particles: they work too, but of course you need a UPLC to get the benefit.
There are quite a few ways to use solid-core. If you've got old HPLCs that can't handle high pressure, they're ideal. If you like 5u pressure but need a bit more resolution a 5u solid-core will give resolution similar to a 3u particle size on almost any LC system, at low pressure. You can also use, for example, 3u solid core to have a go at UPLC-like separation on a conventional pressure system, but UPLC-narrow peaks do depend on a short and dead-volume-free fluid path too, so you won't get really good chromatography from an old, large system just by putting a modern column on it.