Agilent 5975C sensitivity drop off
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
3 posts
Page 1 of 1
I am new to this discussion group but also new to Agilent MSD's. A lab we collaborate with has an Agilent 5975C which does not hold sensitivity for very long. Sometimes after one injection, sometimes after several injections through the night. Agilent have replaced the whole analyser assembly, the electronics, the multiplier three times and it still does the same thing. We have read that the Pfeiffer roughing pumps Agilent uses are not very good. The client also uses the column back-flush technique and I observed this last week. The instrument vacuum drops to 10e-4 torr when the back-flush is being deployed at the end of a sample run. Having worked on high-end magnetic sector instruments since 1985, this seems bizarre. That much drop in vacuum points to a lack of pumping ability, or capacity, to me. I wonder if this is contributing to the fall off in sensitivity. Fall off is also noted even without using the back-flush facility. Would a poor Pfeiffer pump cause this problem?
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
It almost sounds like they have the plumbing wrong on the backflush setup, so that it is sending the flow to the MS not back to the inlet. But if the loss in sensitivity persists even when not using the backflush that should not be what is causing the problem unless the change in vacuum is happening even when the backflush is not operating.
What column are they using and what are they injecting? If for some reason they are exceeding the max column temp it could be bleeding off stationary phase that is killing the sensitivity. Had a bad column once that did that even without exceeding the temperature limit, but found out what it was because the company issued a recall on the lot number for that column.
If vacuum varies over time then that could be the cause of loss of sensitivity, if it is constant then something else is the problem.
Split or Splitless injection? Inlet problems can also cause these symptoms. If you go into manual tune and allow it to scan for several hours with the tune gas does the sensitivity stabilize after a few minutes or does it drop continuously for hours? If scanning in manual tune gives consistent counts over a few hours time, then begin looking at the injector as the potential problem.
What column are they using and what are they injecting? If for some reason they are exceeding the max column temp it could be bleeding off stationary phase that is killing the sensitivity. Had a bad column once that did that even without exceeding the temperature limit, but found out what it was because the company issued a recall on the lot number for that column.
If vacuum varies over time then that could be the cause of loss of sensitivity, if it is constant then something else is the problem.
Split or Splitless injection? Inlet problems can also cause these symptoms. If you go into manual tune and allow it to scan for several hours with the tune gas does the sensitivity stabilize after a few minutes or does it drop continuously for hours? If scanning in manual tune gives consistent counts over a few hours time, then begin looking at the injector as the potential problem.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 27
- Joined: Tue Mar 10, 2009 4:18 pm
Geoff
In case you still have the problem, I will comment on some issue I have seen.
Depending how is configured the backflush accessory, it can happens that a leak at the siltite ferrule brings air directly to the ion source, diluting your sample.
In case you still have the problem, I will comment on some issue I have seen.
Depending how is configured the backflush accessory, it can happens that a leak at the siltite ferrule brings air directly to the ion source, diluting your sample.
3 posts
Page 1 of 1
Who is online
In total there are 137 users online :: 1 registered, 0 hidden and 136 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot] and 136 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot] and 136 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science