Styrene-divinylbenzene columns
Posted: Thu Jun 06, 2013 10:52 pm
Hi LC forum experts
We are performing testing of Doxycycline Hyclate related substances API and tablets as per the EP.
The column specified is 25 cm x 4.6 mm, 8um styrene-divinylbenzene copolymer, we are using an Agilent PLRP-S 100A 8um, part no. PL1512-5800.
The mobile phase is made up of 60 g 2-methyl-2-propanol, 400mL of pH buffer 8 (0.2M potassium dihydrogen phosphate, adjusted with NaOH), 50 mL 1% tetrabutylammonium hydrogen sulphate (adjusted to pH 8 with NaOH), 10 mL of 4% disodium edetate (adjusted to pH 8 with NaOH), made to 1 L with purified H2O. Then vaccum filtered through 0.45 um HVLP membrane filter.
HPLC conditions
1 mL/min
254 nm
60 C column temp
20 uL injection
We are using a Dionex RSLC to run this testing.
The chromatography is good with a new column but, even after only one run the plate count and resolution deteriorate significanly, sometimes the coulmn will last only 4 - 5 runs before system suit can no longer be acheived. The blue chromatogram is from the second run using this column, the plate count of the Doxy peak has gone from 1117 to 539, and the resolution of Impurity A has gone from 2.8 to 1.3.

After running the column is washed using ACN/H2O from a low % ACN upto 90 % ACN for storage.
Does anyone else have experiance with these coulmns and can suggest a way to regenerate or wash them to extend the life that we are getting.
From what I have read these columns should be robust when it comes to these high pH high temperature applications, but the pores of the particles may be getting blocked. All mobile phase and samples are filtered through 0.45 um, is a finer filter required?
I have tried washing the column in reverse but it doesn't seem to improve the chromatography?
Any suggestion much appreciated as we are going through columns at a fast rate
We are performing testing of Doxycycline Hyclate related substances API and tablets as per the EP.
The column specified is 25 cm x 4.6 mm, 8um styrene-divinylbenzene copolymer, we are using an Agilent PLRP-S 100A 8um, part no. PL1512-5800.
The mobile phase is made up of 60 g 2-methyl-2-propanol, 400mL of pH buffer 8 (0.2M potassium dihydrogen phosphate, adjusted with NaOH), 50 mL 1% tetrabutylammonium hydrogen sulphate (adjusted to pH 8 with NaOH), 10 mL of 4% disodium edetate (adjusted to pH 8 with NaOH), made to 1 L with purified H2O. Then vaccum filtered through 0.45 um HVLP membrane filter.
HPLC conditions
1 mL/min
254 nm
60 C column temp
20 uL injection
We are using a Dionex RSLC to run this testing.
The chromatography is good with a new column but, even after only one run the plate count and resolution deteriorate significanly, sometimes the coulmn will last only 4 - 5 runs before system suit can no longer be acheived. The blue chromatogram is from the second run using this column, the plate count of the Doxy peak has gone from 1117 to 539, and the resolution of Impurity A has gone from 2.8 to 1.3.

After running the column is washed using ACN/H2O from a low % ACN upto 90 % ACN for storage.
Does anyone else have experiance with these coulmns and can suggest a way to regenerate or wash them to extend the life that we are getting.
From what I have read these columns should be robust when it comes to these high pH high temperature applications, but the pores of the particles may be getting blocked. All mobile phase and samples are filtered through 0.45 um, is a finer filter required?
I have tried washing the column in reverse but it doesn't seem to improve the chromatography?
Any suggestion much appreciated as we are going through columns at a fast rate