in HPLC calibration wave length accuracy

how i determine wave length accuracy of detector at lower uv wave lenghts i.e. 210 nm , 205 nm
You don't. Almost any signal you see in that region is "end absorbance" which continues to increase as you go to shorter wavelengths (i.e., hard to find anything that has a true absorbance maximum in that region).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
I sort-of feel that wavelength calibration is one of the less useful calibrations we do anyway, at least in quantitative work.

Most things have fairly broad maxima, so if you're out by a few nm, it won't make any difference. And in any case, if the standards and samples are both measured at the same wavelength, the actual exact wavelength used probably doesn't matter. If you're comparing wavelength maxima to expected values as part of a strategy for identifying things, that of course is a whole different situation.
3 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry