Just to be completely clear - a 5 "ppt" (presumably ng/l) gives a bigger peak than a 100 ppt standard ? or is it just that the peaks are bigger than they are expected to be for some of the standards ?
The smallest peak that you can expect to see with an FID is about 0.1 ng, let me guess at a sample volume of 10 ml and assume 100 % recovery into the SPME fiber. Your lowest standard contains only 50 pg (= 0.05 ng) of analyte, so whatever peak you are seeing with the FID it is not your analyte. Your highest standard will give you a clearly detectable peak if you have 100 % extraction to the SPME fiber.
What is your dilution scheme when making up standards, what are the various materials dissolved in an diluted with at each stage ?
Peter
Yes, 5ppt is ng/L and yes my samples are 10ml.
Standards made in MEOH, I start off with a stock of 100 ng/uL of MIB and geiosmin (in MEOH), add 1ul of this to 1mL of MEOH for my internal standards and calibration. From here I make my 5ppt, 10ppt and so forth till I get to 100ppt to complete my cal curve.
I have used this instrument in the past briefly and got great resoluion at the 5ppt level, I always thought that it was my compounds since when I overlay all my peaks in different concentrations, there use to be spot on in the same Retention time etc...
Now, I have having this crazy issue that the 5ppt gives me a taller, bigger area count than my other standards.
Over night I ran another set of curves using the "original" flame tip (I had replaced it previously due to the ferrule disintegrating) and things look better. I have no idea why the flame tip would matter. But I am running some more standards changing gas ratios to 20(He), 40(H2), 350(air) see if this helps.
Thanks for the idea sharing. It helps me greatly to "talk" to someone about it!