What does SPE c18 do?
could you please give me a SPE C18 extraction protocol?

Thanks
In a nutshell, it is used to either extract moderately polar to non polar compounds from an aqueous matrix or it can be used as a filter to remove contaminants from an extract.

We have several applications on our website as do the other SPE manufacturers.

Don
Don Shelly
Don Shelly Consulting, LLC
don.shelly@donshellyconsulting.com
thanks.

Could you please see whether this protocol makes senses?

I'm using SPE C18 (500 mg, 3 ml Column volumn).

1. Activate the column by using 2.5 ml 100% MeOH
2. Prewash the column with 8 ml 0.1% TFA pH 5.0
3. Load the sample (500 ml of serum + 1.5 ml of 0.1% TFA pH 5.0)
4. Wash the column with 2 ml of water
5. elute with 1.5 ml 25%MeOH in water

Can I increase the percentage of MeOH from 25% to 50%? Because I have to dry the samples after the elution step
thanks.

Could you please see whether this protocol makes senses?

I'm using SPE C18 (500 mg, 3 ml Column volumn).

1. Activate the column by using 2.5 ml 100% MeOH
2. Prewash the column with 8 ml 0.1% TFA pH 5.0
3. Load the sample (500 ml of serum + 1.5 ml of 0.1% TFA pH 5.0)
4. Wash the column with 2 ml of water
5. elute with 1.5 ml 25%MeOH in water

Can I increase the percentage of MeOH from 25% to 50%? Because I have to dry the samples after the elution step

I can't think of a good reason to use water in this elution solvent. Water will aid in the elution of polar compounds from a polar sorbent. It will interfere in the elution of non polar compounds from a non polar sorbent. I recommend elution with 100% MeOH. I'm saying this without knowledge of your analytes.

Don
Don Shelly
Don Shelly Consulting, LLC
don.shelly@donshellyconsulting.com
Thanks.
So you think I should use 100%MeOH instead of 25% MeOH, don't you?
yes
Don Shelly
Don Shelly Consulting, LLC
don.shelly@donshellyconsulting.com
It basically adsobs stuff (usually very nonpolar stuff like oils) that would otherwise adsorb to your column. Better to have a bunch of junk stuck to a disposible SPE column than your analytical column. Also the very polar compounds flow right through with a wash.
7 posts Page 1 of 1

Who is online

In total there are 9 users online :: 0 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry