Page 1 of 1
SPE C18
Posted: Thu May 30, 2013 7:33 pm
by panpicha
What does SPE c18 do?
could you please give me a SPE C18 extraction protocol?
Thanks
Re: SPE C18
Posted: Fri May 31, 2013 1:25 am
by Don Shelly
In a nutshell, it is used to either extract moderately polar to non polar compounds from an aqueous matrix or it can be used as a filter to remove contaminants from an extract.
We have several applications on our website as do the other SPE manufacturers.
Don
Re: SPE C18
Posted: Tue Jun 04, 2013 9:32 pm
by panpicha
thanks.
Could you please see whether this protocol makes senses?
I'm using SPE C18 (500 mg, 3 ml Column volumn).
1. Activate the column by using 2.5 ml 100% MeOH
2. Prewash the column with 8 ml 0.1% TFA pH 5.0
3. Load the sample (500 ml of serum + 1.5 ml of 0.1% TFA pH 5.0)
4. Wash the column with 2 ml of water
5. elute with 1.5 ml 25%MeOH in water
Can I increase the percentage of MeOH from 25% to 50%? Because I have to dry the samples after the elution step
Re: SPE C18
Posted: Tue Jun 04, 2013 10:37 pm
by Don Shelly
thanks.
Could you please see whether this protocol makes senses?
I'm using SPE C18 (500 mg, 3 ml Column volumn).
1. Activate the column by using 2.5 ml 100% MeOH
2. Prewash the column with 8 ml 0.1% TFA pH 5.0
3. Load the sample (500 ml of serum + 1.5 ml of 0.1% TFA pH 5.0)
4. Wash the column with 2 ml of water
5. elute with 1.5 ml 25%MeOH in water
Can I increase the percentage of MeOH from 25% to 50%? Because I have to dry the samples after the elution step
I can't think of a good reason to use water in this elution solvent. Water will aid in the elution of polar compounds from a polar sorbent. It will interfere in the elution of non polar compounds from a non polar sorbent. I recommend elution with 100% MeOH. I'm saying this without knowledge of your analytes.
Don
Re: SPE C18
Posted: Tue Jun 04, 2013 10:46 pm
by panpicha
Thanks.
So you think I should use 100%MeOH instead of 25% MeOH, don't you?
Re: SPE C18
Posted: Wed Jun 05, 2013 1:34 pm
by Don Shelly
yes
Re: SPE C18
Posted: Thu Jul 11, 2013 11:57 pm
by MSCHemist
It basically adsobs stuff (usually very nonpolar stuff like oils) that would otherwise adsorb to your column. Better to have a bunch of junk stuck to a disposible SPE column than your analytical column. Also the very polar compounds flow right through with a wash.