my GC-MS (5973) has shown high N2, the ratio to O2 is over 10, somtimes, N2 becomes the base peak after DFTPP tuning. At least no apparent leak as I can see. is this a carrier gas issue (UHP helium) or still a leak in somewhere? Anyone can tell me?
-
- Posts: 960
- Joined: Thu Apr 18, 2013 2:10 pm
Have you tried a different tank of carrier gas? I had this issue a while back and it turned out that the tank was contaminated with N2.
-
- Posts: 350
- Joined: Tue Jun 28, 2011 7:30 pm
It is worth checking if it looks the same after a standard autotune is run; I have seen odd tunes over emphasize the gases on occasion - though N2 as the base peak is pretty out there. I believe the air and water check utility will ask if you want to reset to standard conditions (or some such phrasing) for the same reason.
-
- Posts: 69
- Joined: Mon Jul 11, 2011 3:15 pm
We had this problem last year, switched tanks which didn't help, went to change the trap and discovered one of the swagelock fittings had been installed incorrectly (for years?!), don't know why it showed up suddenly. Also thought it was curious that it was just nitrogen that showed up, little water/oxygen.
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Is your Helium cylinder below 500psi? Sometimes when it gets near empty you will find that N2 has settled to the bottom of the cylinder.
Also take some canned air which is a fluorocarbon and spray it around your side plate seal and see if it shows up when scanning with the PFTBA valve closed. It is really easy to get a small piece of lint on the o-ring and have good vacuum but still have a leak large enough to see N2 as the base peak.
Also take some canned air which is a fluorocarbon and spray it around your side plate seal and see if it shows up when scanning with the PFTBA valve closed. It is really easy to get a small piece of lint on the o-ring and have good vacuum but still have a leak large enough to see N2 as the base peak.
The past is there to guide us into the future, not to dwell in.
5 posts
Page 1 of 1
Who is online
In total there are 26 users online :: 0 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 26 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 26 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science