5973 GC Problem with relative Abundances
Posted: Wed May 29, 2013 5:59 pm
I was hoping that I might be able to get your input on an issue that we are having here at the lab. We have a 5973 GCMS that we are running VOC’s on. The problem we are seeing is a rise in the relative abundances of the ISTD’s throughout the run of the curve.
At first I thought that this issue might be due to varying amounts of MeOH in the vials, but I made standards with the same amount of MeOH and still saw a rise in abundance as the higher concentration ran. For example, our 5ppb standard will run with area counts around 500K, 1Million, 1.1Million, 7200K. By the 200ppb standard exhibited 820K, 1.4Million, 1.6 Million, 1.1Million
We run with the following configuration:
4552 OI (Archon) Autosample
OI Eclipse concentrator (also tried with a 4560 Concentrator)
6890 GC Oven
5973 Network Mass Spec
I have tried the following in an attempt to distinguish where the problem might be coming from.
1) I replaced the concentrator and the XFER line to the GC with no change, under the assumption that there could have been active sites. Still saw the ISTD rise over the course of the curve.
2) Ran water samples instead of Soil samples (both directly through the concentrator and using the autosampler) and had the same results of an increase in the ISTD throughout the run.
3) I have replaced the XFER line from the Autosampler to the Concentrator. No change
4) I have replaced the Concentrator and still no change.
5) Replaced the 1uL ISTD loop
6) We have tried multiple traps
7) Direct injection: I did a direct injection of standards to represent a low point (5ppb) midpoint (25ppb) and high point (200ppb) and so no change in the ISTD area counts. They all remained constant 500K, 1M, 1M, 500K.
Spiked directly three points directly onto the trap (5ppb, 50ppb, 200ppb)
These problems have all arisen since we had maintenance on the instrument in which the following were replaced. The parts were:
Clean the source, replaced Electron Multiplier, Replaced Repeller, replaced DO, and replaced Repeller Insulators.
Any help would be much appreciated
At first I thought that this issue might be due to varying amounts of MeOH in the vials, but I made standards with the same amount of MeOH and still saw a rise in abundance as the higher concentration ran. For example, our 5ppb standard will run with area counts around 500K, 1Million, 1.1Million, 7200K. By the 200ppb standard exhibited 820K, 1.4Million, 1.6 Million, 1.1Million
We run with the following configuration:
4552 OI (Archon) Autosample
OI Eclipse concentrator (also tried with a 4560 Concentrator)
6890 GC Oven
5973 Network Mass Spec
I have tried the following in an attempt to distinguish where the problem might be coming from.
1) I replaced the concentrator and the XFER line to the GC with no change, under the assumption that there could have been active sites. Still saw the ISTD rise over the course of the curve.
2) Ran water samples instead of Soil samples (both directly through the concentrator and using the autosampler) and had the same results of an increase in the ISTD throughout the run.
3) I have replaced the XFER line from the Autosampler to the Concentrator. No change
4) I have replaced the Concentrator and still no change.
5) Replaced the 1uL ISTD loop
6) We have tried multiple traps
7) Direct injection: I did a direct injection of standards to represent a low point (5ppb) midpoint (25ppb) and high point (200ppb) and so no change in the ISTD area counts. They all remained constant 500K, 1M, 1M, 500K.
Spiked directly three points directly onto the trap (5ppb, 50ppb, 200ppb)
These problems have all arisen since we had maintenance on the instrument in which the following were replaced. The parts were:
Clean the source, replaced Electron Multiplier, Replaced Repeller, replaced DO, and replaced Repeller Insulators.
Any help would be much appreciated