SJ.LI » Thu May 30, 2013 6:35 am
SJ,
You need to give a bit more information. If you are presently using two valves on the GC, one to inject the sample and the other to isolate the molecular sieve column, then you can use the same columns for the compounds in your message. You may have to change the valve switching times. If there are other compounds in your sample that you are not interested in measuring, then there could be problems.
Gasman
Gasman,
I am sorry for the insuffient information i given.
Actually, my lab is using the GC-TCD system (Agilent 6890N) for the Methane and CO2 analysis.
There is individual injector, column and detector for each channel.
Channel "A"
Parameters: Methane and O2
Column: HP-MOLSIV (30m 0.32 25)
Detector: TCD
Channel "B:
Parameters: CO2
Column: HP-PLOT U (30m 0.32 10)
Detector: TCD
Both of the two channels are using Helium as the carrier gas.
If i want to develop a new method for N2, Ethane, Propane, n-butane and CO, are the presently columns suitable for the parameters? or should i use another carrier gas ?
SJ