Posted: Wed May 29, 2013 12:31 am
We are looking at using the Restek EZ No-Vent System with our Agilent GC(6890N)-MS(5975B) so we do not need to vent the MS when changing columns. In the past we have used a microfluidic switch, but found that it greatly reduced the sensitivity, which was a problem for trace analyses of environmental contaminants.
Does anyone have experience with this system and have you found any major and/or recurring issues (e.g. decreased sensitivity). I have seen it mentioned somewhere that the ferrules get stuck in the EZ no-vent and it needs to be replaced regularly. Has anyone else experienced this?
Finally, our system uses EPC. Has anyone had any issues setting up the EZ no-vent with EPC software (which cannot compensate for the changes in flow) - or is this only an issue for software that uses retention time locking.
Thanks.