Hello,
I guess that what you would choose as a column-cleaning solvent will kind of depend on what you are chromatographing. Methanol already is a pretty strong normal-phase eluent--think of it in the opposite terms of reversed-phase eluents. THF, IPA, IPA/chloroform or hexane are good for cleaning reversed-phase style modified silica--these are non-polar relative to water, ACN or MeOH, roughly speaking.
MeOH is far more polar than typically-used normal-phase solvents such as hexane, methylene chloride or MTBE, so in a fashion, MeOH would be a good strong eluent for cleaning. Water might be stronger still for polar analytes, but you don't want to mess with water levels on bare silica...whatever the case, make certain that whatever you choose to clean with is completely miscible with MeOH...your eluent of choice. IPA and acetone are great for converting between solvents of different polarity, should it come to that kind of thing...of the two, I've had better luck with IPA (acetone, depending on source, can have a lot of "junk" in it, even if it's supposed to be HPLC-Grade.)
Good Luck!! See what others may think. Tom's a faster typist than I am--and I agree with his comments. Tom, please pardon my tailgating!