Mant0 and Tom Jupille,
I am thoroughly embarrassed...I allowed the scale of the instrument to "throw" me. After sitting down and doing my own set of calculations I'm in agreement with Tom. Basics will always be basics, though, which I somehow forgot for a bit--probably was carelessness on my part.
A column-volume for a 0.075 mm x 150 mm column, leaving out stationary phase "bulk," is ca. 0.66 uL (and I missed this above and apologize to mant0). If the eluent flow was 0.2 uL/min, 3 min of run-time would pass .6 uL of eluent...approximately one column-volume...another way to say this is what Tom said, no retention of analyte(s). It's as if there was no stationary phase in the column...throwing in the porosity of the stationary phase (unless the S.P. is non-porous silica like Kovasil) the "blob" k' is still only slightly greater than one...no real "chromatography" is being done.
Either there is a massive leak between the injector and column inlet, or there is a gigantic channel in the column's stationary phase.
During the course of the evening I came across a cool book:
More Practical Problem Solving in HPLC Author is Stavros Kromidas, ISBN: 978-3-527-31113-2, Paperback, 309 pages, January 2005, Wiley VCH
Chapter 2.2 is pretty much excellent reading...maximum injection volume suggested for a column of approximately the same dimensions as the one you are using is 30 nL (some downscaling for an analytical scale guy like me!) and the column mass load suggested is 6 ng (also quite small to me). Suggested flow rate for eluent was 350 nL/min. It's good to live by a good Library
I apologize Mant0, I think, but this was kind of a fun problem for me to help with...very educational. Hope this Helps...my thanks for listening to my opinions, and Best of Luck again!
Learning Doesn't Ever Really Stop...and Tom, my thanks and appreciation for coming to our rescue!