I'm using an HP6890 GC for 8260 Volatiles analysis and the instrument is having a problem holding a constant head pressure. I'm running @ 1.0mL/min, 36cm/sec and have a head pressure of 7.4 on the GC with a total flow of 23.6. The pressure fluctuates from 7.2 to 7.6 with a setpoint of 7.4 with a constant oven temp of 50'C and inlet temp of 230'C and still has the fluctuation with an increase in head pressure as the oven temperature increases. I did notice the the head pressure will remain constant when I decrease the inlet temperature (stayed at a constant pressure with no fluctuation when the inlet was less than 150'C) but started fluctuating again once the inlet temp was raised back to 230'C. Agilent suggested that I clean the split vent (running in splitless) and change the copper tubing and chemical trap but it seems more like a weldment issue now (using a fliptop weldment and merlin septa). Any thoughts? Thanks for your help.