Does Viton O-ring withstand high temperatures?
Posted: Thu May 23, 2013 1:44 pm
by hplc.chem
Hi all,
I am strugling with high background peaks problem. I am currently suspecting the Viton O-ring I use for the GC inlet liner.
I am using Varian GC-MS, with 1177 Injector. I noticed that the O-ring became flat! I am using the Injector above 300 oC.
Do you think the Viton O-ring withstand such a high temperature?
I checked in the Agilent website for the Max temp for this O-ring but I failed to find a figure.
Any help in this regard is highly apprciated.
Thanks in advance.
Re: Does Viton O-ring withstand high temperatures?
Posted: Thu May 23, 2013 2:59 pm
by Steve Reimer
Without knowing the supplier it is a little tough to get precise information but for generic Viton o-ring specs try:
http://o-rings.com/pdf/TRV70875.pdf.
I have used these for a sealing application in a GC column oven. I found that they worked until I started running PBDE analysis with a 20 minute hold at ~320 degrees. They deformed enough to leak within a few hours. With a normal PCB analysis and only a short hold at that temperature they were good for weeks.
Posted: Fri May 24, 2013 3:30 pm
by Yama001
The vitons will deform at those temps, they just need to be replaced more often is all.
Agilent sells a graphite o-ring for 350 and above - I do not like to use them.
Posted: Fri May 24, 2013 4:47 pm
by hplc.chem
Thanks Steve and Yama,
I actually find the ring FLAT after using the high temperature above 300. But is there any possibility for bleeding to the injector ad then to the column?
Yama: what is the problem with the graphite O-rings?
Posted: Fri May 24, 2013 6:29 pm
by Yama001
I run with viton rings at 300 - they flatten out but maintain a seal - I do replace them every week or so - before they get ragged edges.
I don't like the graphite ones because careless operators tend to smear graphite around is all - they are pretty reliable otherwise.
Posted: Sat May 25, 2013 1:11 am
by Don_Hilton
I am a fan of the graphite o-rings - having had to peel the remains of a viton o-ring off the inlet too many times to please me.