negative peak used to determine void volume
Posted: Sat May 18, 2013 9:53 am
Hi guys,
In most of the methods I used so far, void volume is determined by injection of NaNO3 solution, and the retention time of NaNO3 (To) is used to calculate k´ of the analyte.
In one method for Omeprazole assay that I started to use recently, a negative peak (at about 1 minute) from injection of dissolving medium (blank) is used to determine k´ of Omeprazole.
I would like to understand what causes that negative peak and why it can be used for determination of To (or void volume Vo)
Mobile phase: 26% acetonitrile, 71% phosphate buffer pH 7.4, 3% of TBAHSO4
Dissolving medium: 20% acetonitrile, 80% phosphate buffer pH 11.0, I =0.01 (that is the blank solution that has the negative peak at RT about 1 minute)
Flow rate: 1 ml/min
Column: Microspher C18, 3um particles, 4,6*100 mm
Detection: UV-vis detector, wavelength 280 nm
What causes that negative peak and why the retention time of that peak can be used as To ?
Almin
In most of the methods I used so far, void volume is determined by injection of NaNO3 solution, and the retention time of NaNO3 (To) is used to calculate k´ of the analyte.
In one method for Omeprazole assay that I started to use recently, a negative peak (at about 1 minute) from injection of dissolving medium (blank) is used to determine k´ of Omeprazole.
I would like to understand what causes that negative peak and why it can be used for determination of To (or void volume Vo)
Mobile phase: 26% acetonitrile, 71% phosphate buffer pH 7.4, 3% of TBAHSO4
Dissolving medium: 20% acetonitrile, 80% phosphate buffer pH 11.0, I =0.01 (that is the blank solution that has the negative peak at RT about 1 minute)
Flow rate: 1 ml/min
Column: Microspher C18, 3um particles, 4,6*100 mm
Detection: UV-vis detector, wavelength 280 nm
What causes that negative peak and why the retention time of that peak can be used as To ?
Almin