Calibration table vs. Reprocessed sequence

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
I have a question: When you calibrate on different levels using the ISTD amount in the calibration table, why do you have to write this amount also in the Sequence table (before reprocess)? How does this action affect your results?

Thanks.
Hello

It is for automation purpose...let say you're controlling process and you want to stop sequence when sample if out of spec (out of range). If you have amount (concentration) set up in sequence table system can calculate results during your run.
Second reason could be audit trail - if you have concentration/amount specified before your run you can not adjust results after.

You don't have to set it up in sequence table because you can always process results after.

It is only my opinion :)

Regards

Tomasz Kubowicz
Thank you for your answer!

The ISTD amount is entered after the run, in the evaluation of results step (in Offline modus of Chemstation). I don't know if this action affects my results in calibration or it is just for the Audit Trail reason.
3 posts Page 1 of 1

Who is online

In total there are 35 users online :: 0 registered, 0 hidden and 35 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 35 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry