- Posts: 95
- Joined: Fri Nov 30, 2012 5:18 pm
Column ODS HYPERSIL (mm) 100x4.6, 3µm
Buffer pH 8.0 3.25g Tetra-n-butylammonium hydroxide, 40%w/w +
1.3g di ammonium hydrogen phosphate + 900mL water, pH adjusted into 8.0 with dilute phosphoric acid (afterwards dilute to volume into 1000mL with water)
Mobile phase Mobile Phase A: Buffer pH 8.0
Mobile Phase B: 100% Methanol (HPLC Grade)
Gradient Program
Time (minutes) %A(v/v) %B(v/v)
2 45 55
3 40 60
13 35 65
20 12 88
26 12 88
27 0 100
31 0 100
33 45 55
Flowrate 1.6 mL/min
Injection Volume 20 µL
UV Detection 286 nm
Column Temperature 40°C
I used the above mentioned hplc condition for the analysis of etofenamate's relate substance. At first, I was getting good result but as time goes by, my active's retention time changes dramatically from 11-12 minutes to 15-17 minutes. The sample formulation we had was gel type. Following the procedure below for sample preparation, we experienced peak shouldering. Apart from that, we experienced also prime overpressure, compromising the whole trial run. What could be the reason from the problems we faced lately. thanks.
SAMPLE PREPARATION: 1000mg of sample into 50mL vol. flask, diluted to volume with 70% methanol then sonicated for 30 minutes to dissolve. Sample was filtered with 0.45µm nylon filter prior to vialing and injection.