Hello
I'm running a method which measures derivatised olive oil by using an internal standard ratioed to the C16 & C18 peaks. Now I'm about to inject samples that will have much larger results than normal (at least 2x) so I’m thinking of either injecting less sample or increasing the split at the injection to keep the peaks within the max voltage reading on the FID. I normally inject 1µl and have a split ratio of 50:1. Which would be more accurate. I would guess injecting less but is there a rule of thumb/calculation to follow with this sort dilemma?
I'm running a method which measures derivatised olive oil by using an internal standard ratioed to the C16 & C18 peaks. Now I'm about to inject samples that will have much larger results than normal (at least 2x) so I’m thinking of either injecting less sample or increasing the split at the injection to keep the peaks within the max voltage reading on the FID. I normally inject 1µl and have a split ratio of 50:1. Which would be more accurate. I would guess injecting less but is there a rule of thumb/calculation to follow with this sort dilemma?