All,

Our Waters 2487 dual absorbance detector is turning on and goes through the warm up but then shut off right when it gets to the "homing optical filter" verification stage, and then it turns back on and repeats the cycle.

I've changed out the lamp but it's still behaving the same way. Would appreciate it if anyone could offer some suggestions on what's causing this issue, in particular, if you've had similar experiences with the detector.

Would appreciate it. Thanks.
Hi,

The issue is with your LVPS unit. It needs a new Low voltage power supply (LVPS). Contact your local Waters service department or who ever looks after your service.

Regards,
Bull76.
This has been discussed before. This link should help you out.

viewtopic.php?f=1&t=14128&hilit=2487+issue
3 posts Page 1 of 1

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry