I'm in process of developing methods for the purification of a variety of peptides, of which some display low solubility in water. In these cases I end up with very dilute samples and in turn I have to inject a larger volume/run.
Does the sample volume have any effect on the resolution?

In advance thanks
Erik

Re: Effect of Sample Volume

Posted: Tue May 07, 2013 8:20 am
by HPLCaddict
Clear answer: It depends :D

Generally, yes. Larger sample volumes mean that your sample reaches the column in a larger solvent "plug". The larger the injection volume, the wider the peak will be. Wider peaks means less resolution.
And now the BUT: If your sample solvent is weaker (meaning less elution power) than the mobile phase, the sample will get concentrated on the column head as it reaches the column. This is especially pronounced with gradient separations, where the initial mobile phase is too weak to significantly move the sample. Therefore, with gradient separations you might be able to actually inject rather huge sample amounts without significant peak broadening if the sample solvent is weaker than the initial mobile phase.
In the end, you have to try. Occurence of peak broadening or distortion because of a too large injection volume is a complex function of solvent and mobile phase composition, column dimensions, flow rate, injection volume (obviously :D ) and even the HPLC itself (tubing dimensions etc.).

Re: Effect of Sample Volume

Posted: Tue May 07, 2013 12:11 pm
by holmerz
Thanks for the answer.

Typically organic modifiers (Acetonitrile) aids the dissolution of the peptides, so a scenario with less organic modifier than the starting gradient, is rare, at least in my case.