Waters 717 Leaks during Needle Wash

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
I think my title says it all!

I've tried swapping and calibrating 3 different seal packs, each being brand new and unsuccessful. I just need this to stop leaking when the needle wash happens during a run.

I'm completely stumped. Any suggestions?
Josh

Metrology Technician
Boston Analytical
1. Valve 4 could be broken, I dont know if it is part of a seal pack.
2. Seal wash bottle should be on the bench, not above the AS (this has been know to cause leaking)