Configure Agilent 1100 ALS to wait for injection signal

Posted: Sun May 05, 2013 1:04 pm
by isovitexin
Hello,

We have in our lab a 1100 LC system and we want to conect to a Metrohm Bioscan detector, I have been reading and it seems that the easiest way is to control the Bioscan in the ICNet interface and record the analog output of the detector and send it to Chemstation, but for this we would need the Agilent A/D box that we do not have.

Other option, more cheaper, would be to conect the ALS (autosampler) to the Bioscan using the BCD cables for the ALS (that we already have), but Bioscan system does not expect a start signal, it wants to send it. Does anyone know if you can configure the ALS in the Chemstation for waiting for an input signal to start the run?? I suspect that the ALS would take different times to inject but I think I can compensate that using an injection program.

Thanks in advance for the help!!

Re: Configure Agilent 1100 ALS to wait for injection signal

Posted: Wed May 08, 2013 5:50 pm
by sepscientologist
You can use the remote out cable. It should work like this but might need some experimentation.
It is a DB9 pin I believe, and driving pin 2 ttl low should cause the autosampler to prepare to run
and then driving pin 3 low will start the injection

Re: Configure Agilent 1100 ALS to wait for injection signal

Posted: Wed May 08, 2013 5:58 pm
by isovitexin
Many thanks!!! :D

We can try that configuration I think we have already that DB9 cable (Agilent remote cable), but I suppose that some configurations should be done in Chemstation (because ALS would not wait for signal for start and inject the sequence), Manual is really useless for this topic.

Thanks again for the help!