Straight line doesn't pass through origin and intersect x-axis at 225 ppm. R square value is 0.9986. I am using ELSD detector Base line was at zero and blank run was also at zero. I am using bracketing standards i.e. my sample concentration is in between my standard concentration. I prepared my standards by myself after dilution of sample and couldn't find pure standards. Retention time is 8 min. What can be possible reason? and is this acceptable?
- how far off are you? If the origin is within the prediction interval, then technically, your line *does* pass through the origin.
- how likely are you to get samples that are down near LOQ? If all of your samples are in the middle of the linear range, and you always run bracketing standards, then whether the calibration passes through the origin is relatively unimportant.
- if you analyzing over a wide range of concentrations (e.g. several orders of magnitude), then consider doing a weighted least-squares fit (or using a log-log fit) to compensate for the heteroscedasticity of chromatographic data.