when you try to clean up the inlet or outlet valves, the needle seat, parts of the pump etc and you do sonication, what is the better to use? H2O mQ, MeOH, Isopropanol, H2O with a little acid(glacial acetic)?
Is there any harm that can be made?
As a general rule, something that's a good solvent for whatever kind of junk is likely to have accumulated. Water with a bit of methanol would be a good choice for salt crystals, isopropanol might be a better choice for residual surfactant (e.g., ion-pair reagents).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Be careful. Sonication may damage an intact check valve.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
3 posts Page 1 of 1

Who is online

In total there are 21 users online :: 2 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry