RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Sat Apr 27, 2013 2:40 pm
by YAG
Dear friends,
I am working on one method but having problem to execute it and meeting system suitability
Method is
Column:- DB-1701 30m X 0.25mm 1.micron
temp:- 50 Deg 5 min - 10 Deg/min -200 deg
Detector Temp:- FID - 250 Deg
Injec Temp:- 25 Deg
Gas :- Helium @ 10 psi
Inj Volu :- 3 microlit

Problem :- Unable to achieve RSD lmit 2.0 % for 5 std Injections.

Please help me

Re: RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Sat Apr 27, 2013 3:47 pm
by dblux_
what solvent, why so low injector temperature ?

isn't your column overloaded ?

what mode of injection ?

Re: RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Sat Apr 27, 2013 8:46 pm
by chromatographer1
Your problem could be as simple as the type of injection liner or the placement of the column in the injector.

Much more information is needed before anyone can perhaps determine the issues involved.

best wishes,

Rod

Re: RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Sun Apr 28, 2013 8:14 am
by YAG
what solvent, why so low injector temperature ?

isn't your column overloaded ?

what mode of injection ?

Solvent :- Ethyl Alcohol
Yes! column is overloaded due to sample injection.
Mode of Injection is Split - split ratio is 100:1 on Thermo Focus

Re: RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Sun Apr 28, 2013 8:16 am
by YAG
Your problem could be as simple as the type of injection liner or the placement of the column in the injector.

Much more information is needed before anyone can perhaps determine the issues involved.

best wishes,

Rod

Dear Rod
Split Liner 5 mm and the distance is 4 cm in Injector port

Please help

Thanks

Re: RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Sun Apr 28, 2013 8:48 am
by dblux_
Solvent :- Ethyl Alcohol
Yes! column is overloaded due to sample injection.
Mode of Injection is Split - split ratio is 100:1 on Thermo Focus
In my opinion you have to avoid column overloading by analytes (not solvent).

I don't know what is "Thermo Focus" - is it a kind of PTV ?

If it is common split inlet than your inlet temperature is substantialy to low.

Re: RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Sun Apr 28, 2013 3:56 pm
by chromatographer1
" If it is common split inlet than your inlet temperature is substantialy to low. "

I agree. Try 80C as a minimum and go up from there.

Rod

Re: RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Sun Apr 28, 2013 6:14 pm
by dblux_
BTW - you split your sample 100:1 so probably column is not overloaded - depends on concentration of analytes.

Re: RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Mon Apr 29, 2013 6:38 am
by Peter Apps
The very low inlet temperature has already been commented on - to get that low you must have cryogenics in the inlet, which suggests that it is a PTV of some kind, but if it is you need to programme its temperature after the injection.

Why inject 3 ul and then split 100:1, why not inject 1 ul and split 30:1 ?

Peter

Re: RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Mon Apr 29, 2013 8:33 am
by dblux_
Check condition of autosampler syringe.
Are there bubbles in syringe when drawing samples ?

1 µL injections will be more reproducible than 3 µL (less bubbles).

Re: RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Tue May 21, 2013 4:27 pm
by YAG
Solvent :- Ethyl Alcohol
Yes! column is overloaded due to sample injection.
Mode of Injection is Split - split ratio is 100:1 on Thermo Focus
In my opinion you have to avoid column overloading by analytes (not solvent).

I don't know what is "Thermo Focus" - is it a kind of PTV ?

If it is common split inlet than your inlet temperature is substantialy to low.


I am realy very sorry , Injector Temp is 250 Deg C and Thermo Focus is make of GC.

Re: RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Tue May 21, 2013 4:32 pm
by YAG
Check condition of autosampler syringe.
Are there bubbles in syringe when drawing samples ?

1 µL injections will be more reproducible than 3 µL (less bubbles).

These actions I already takeup while performing the analysis.
I also tried to inject 1 µL Injection but required limit is still not achieved.
Is there any other problem? Please help me.

Re: RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Tue May 21, 2013 4:42 pm
by chromatographer1
I suspect you are trying to inject too much sample, or your column is misalligned in the injector. 2% RSD for split injections is about the norm.

What is your split ratio? Rather than injecting 3 microliters, why not do splitless?

You really need to tell us more about the sample. What is the type of liner beside ID, (there are many kinds) ?

good luck,

Rod

Re: RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Wed May 22, 2013 7:03 am
by Peter Apps
Injecting 3 ul of ethanol to a hot inlet gives a very large vapour cloud, a pressure pulse and disrupted flow and split ratio. As suggested before; decrease you injection volume to 1 ul and change you split ratio to 30:1 (you say that you did 1 ul injections, but did you decrease the split ratio ?).

Do you have to use ethanol as the solvent ?

Peter

Re: RSD Problem on DB-1701 column Inj 3. Microlit

Posted: Thu May 23, 2013 3:40 am
by thohry
I think at normal conditions, injection volumn in a GC is 2.0 microliter max. There maybe a backflush and that could affect the RSD%.