Ghost Peak in Blank

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hi Everyone

I am developing a method for diclofenac on UHPLC-MS/MS (Flexar -15 connected with API Qtrap 5500) using Eclipse plus C18 RRHD column (Agilent Technologies). I was running an isocratic method (methanol:0.1% formic acid::90:10) but now I am trying to develop a gradient method. The gradient is set up with varying % of methanol and 0.1% formic acid in water: 0-0.5 min:0.1%FA (100%), 0.5-1.5 min: the methanol percent changes from 0 to 90%, 1.5-3.5: methanol 90%, 3.5-4.0 min: methanol back to 0 % and 4-5 min: 0.1% FA (100%).

In the isocratic method, the mobile phase injections are clean; however, I see a peak of very high intensity (4.00 e+04 peak height) when I run gradient method. I made several injections (100 injections) of the blank but the peak is consistent in every injection. I tried following needle wash solutions: (a) 50% methanol in water (b) 50% methanol in water with 0.1% formic acid (c) methanol:acetone:water::50:25:25 (d) methanol:acetonitrile:water::50:25:25 (e) DMSO:methanol:water:: 5:50:45 . None of these solution helped to remove the peak. I also removed the needle from autosampler and cleaned with acetonitrile by sonicating for 15 mintues. I washed column with methanol:water (90:10) for 2 hours but this did not alter the intensity of the peak. Even I injected air out of empty vial and this also gave the peak with the same intensity. Any idea what could be the problem and how can I solve it? Any input is greatly appreciated.
Have you tried fresh or even another brand of water for your eluents?
Does the peak have the same m/z as your analyte? If so, you probably have gotten some of it contaminating your system, most likely in the A solvent reservoir.

Run the "three blank gradients" test described in the mini-tutorial on gradient artifact peaks (about 4.5 minute in):
viewtopic.php?f=31&t=19085
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
3 posts Page 1 of 1

Who is online

In total there are 12 users online :: 3 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot], Google Adsense [Bot] and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry