HP 5890 IIA GC- oven and fan not turning on
Posted: Thu Apr 25, 2013 4:38 am
I'm attempting to repair an HP 5890 IIA GC coupled with an HP 5970 MSD and am having quite the major issue with the 5890. The oven and fan do not turn on! A few symptoms:
-the "oven temp" button on the front of the instrument is completely non-responsive.
-timed events run fine, programmed heating sequences progress correctly, no problem with internal clock.
-does not pass the self-check due to me not having installed the detector yet. (WARN: DETECTOR NOT INSTALLED) there is nothing about the oven in these initial error messages.
-the "not ready" light stays on solid because the oven temperature is not ready.
-the fan blades are not frozen; when prodded they freely rotate
-there is no evidence of scorching or heat damage (though i have not taken out the fan to look at the heating element)
-We tested the fuses on the AC power board and they are all fine.
At the moment we suspect that K1 on our AC power board (PCB 05890-60050) may not be working and i would greatly appreciate if anyone had schematics showing electricity flows for this board they'd like to share. And any ideas or comments would be greatly appreciated!