Deactivating active sites with silylation reagent?

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hi all I ma having trouble with my amino acid method with active sites. The method is known to be quite sensitive to active sites. I am noticing a drop and even complete disappearance of the most polar analytes specifically ethyl chloroformate derivitized lysine, histidine, and glutamine. I replaced the septa and liner and tried washing the column with my solvent pump kit (MeCl2 followed by 20%water in MeOH). The separation is fine and the nonpolar amino acids are fine but the polar ones are vanishing. Column is a DB-5MS injection is 10:1 split He with a Agilent 4647 liner.

Would injecting a silylating reagent help?
First, determine if the column is the problem. Install temporarily a new column and see if the problem goes away.

If it does, then try to fix the old column, or throw it away.

best wishes,

Rod
2 posts Page 1 of 1

Who is online

In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry