i broke my micro ecd!!!
Posted: Wed Apr 24, 2013 5:49 pm
i was doing regular PM on my 6890 GC with dual ecd and noticed that a 2" piece of column was broken off inside the detector... i loosened the swage nut and dropped the MUG adaptor out of the detector and that is when my gas line broke right at the body of the adaptor... NUTS!!
i overnighted a new adaptor and replaced it... Leak checked it per literature fixed any leaks and retested...
restored operating conditions and the counts dropped from 840k to 90k pretty fast, it was then that i noticed that the vent line was still stoppered from the leak check so i unstoppered it and reattached the hose that vents it to our air handling system, counts have jumped to 200k and as i write this are slowly creeping past 250k...
what the heck is happening?
why does this happen on the same week the boss gets PT's.
any discussion is appreiciated.
Thankyou
casey
i overnighted a new adaptor and replaced it... Leak checked it per literature fixed any leaks and retested...
restored operating conditions and the counts dropped from 840k to 90k pretty fast, it was then that i noticed that the vent line was still stoppered from the leak check so i unstoppered it and reattached the hose that vents it to our air handling system, counts have jumped to 200k and as i write this are slowly creeping past 250k...
what the heck is happening?
why does this happen on the same week the boss gets PT's.
any discussion is appreiciated.
Thankyou
casey