More recently I have come across a unique small molecule compound. True the sensitivity on our QQQ is there, we still have near 3/1 S/N near 1 ng/mL. The problem is , when running a wide calibration curve (with Internal Standard of course) from say 1ng/mL , 5 , 10, 25, 50, 100, 250, 500, 1000, 2000 ng/mL I observe a loss of linearity at the low end or high end, and the curve looks very quadratic.
It is quite troublesome to see the least after being used to perfectly linear response from other compounds.
I noticed that if i process the curve from 1-100 it is perfectly linear, similarly if I process from 50-2000 it is linear.
Its almost like you need two separate curves, one for the low end and one for the high end.
My question is simple, is it appropriate to run a wide range of standards, then when running an unknown sample for example at higher concentration you can simply drop the lower standards, similarly, if the concentration is found lower, can you drop the higher standards?
Is this practice acceptable? If not, what do you guys suggest? Has anybody come across a similar situation?
I am preparing for a rodent PK study, and probably we need the sensitivity.
I have fiddled with this method quite some time, and found it very hard to increase linearity, probably one negative factor is it takes lots of FV and CE and the compound is a larger small molecule. Also the internal standard is a structural analog and not isotope.