GC-MS automatic run

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

13 posts Page 1 of 1
Hi everyone,
I have a problem with my GC-MS (GC: Agilent 6890, MS: Leco): after each injection (with Combi PAL autosampler) I must run the GC by pressing the "Start" key on the GC panel. How can I automate GC runnig? Is there any setting in the LECO softwere or in GC panel? (actually after each injection, GC reamins in "Pre Run" until manually pressing the "Start" key" ) .

Thanks a lot.
There should be a start connection for the CombiPAL wired at the back of the GC. The LECO also needs a start signal - and I dont remember the details. While I have set up several of these, it's been a few years. It seems to me that the configuration for this is in the LECO manuals. They shoud be in electronic format on the instrument computer.

Was the instrument just set up by a vendor?
The instrument has not been used for 5 years and now I start to work with it. I don't know about setting up history. Everything works good except the auto-run.
Our PAL has an "Interface" connector that connects to the "Remote" jack on the back of the 6890 GC (its a split cable that also splits to the MS - its an Agilent MS so that part may be different for the LECO).
I exactly have the very same problem and I am working on it. Let me know your email and I will let you know if I found the solution.
Problem is solved. Mine is working ok now. What version of Chromatof you are using? Let me know you email and I will guide you through required steps but need some info not to make any risky steps.
My problem is also solved. But I have other problems too:

1- My GC-MS has a FID detector. Can I use it in Chromotof software or it must be run with Agilent Chemstation? If so, how?

2- For sensitivity check, what compound should I use? The Leco manual suggested HCB at 2 ppb, but I haven't it.

Thanks

my e-mail: a.johnm@yahoo.com
If you wish to use the FID along with the Pegasus, you should be able to take that signal into ChromaTOF.

HCB is readily available as a pre-made environmental standard. I typically order a couple of vials at a time. Typically you need to dilute the standard, so one vial will last you a l-o-n-g time. You need to keep the original solution tightly capped.
Dear Walshbig

My problem also exist. I though my problem is solved but I have it again. I was wondering if you could help me.

GC: Agilent 6890
ChromaTOF Version:4.40
TOF version: Pegasus GC-HRT
Combi PAL: from Varian company without Head space and SPME tray (only for solution direct injection).


my e-mail: a.johnm@yahoo.com
I sent you an email. Check your email.
There is a problem with your email. Please check and let me know if it is fixed. See bellow for the yahoo message:

Remote host said: 554 delivery error: dd This user doesn't have a yahoo.com account (a.johnm@yahoo.com) [0] - mta1343.mail.ne1.yahoo.com [BODY]
Dear Walshbig

You are in right. My email was defected. I'd really appreciate if you could mail me again (to my new ID: alcorn.mjohn@yahoo.com).

(problem: data aqusision is run only after manually pressing the "Satrt" button on the GC front pannel)

Thanks a lot.
Check you email
13 posts Page 1 of 1

Who is online

In total there are 176 users online :: 1 registered, 0 hidden and 175 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Amazon [Bot] and 175 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry