Varian GC-3800

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hi everyone

I have a Varian 3800 GC. I set the column flow at 1 ml/min and split ratio at 10 without septum purge, but the software shows a total flow of 24 ml/min ! In other split ration also there is a constant excess flow (~ 13 ml/min).
Surprisingly when the split ratio is set to zero, total flow remains at 21ml/min.

Is there any other flow adjustment in the software? Is this problem related to "gas saver", if so, how should I adjust "gas saver" ?

Thanks a lot.
Your first step needs to be to measure the actual flow out of the split vent, using a flow meter that is independent of the GC flow control - you cannot troubleshoot a system using that system's measurements.

Check that you have the correct column dimensions entered, and the correct carrier gas selected.

Peter
Peter Apps
2 posts Page 1 of 1

Who is online

In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry