Help Lowering Retention Time
Posted: Thu Apr 18, 2013 1:29 am
My boss asked me to develop a HPLC method for three flea killers, PBO(Piperonyl Butoxide), Sumithrin and Nylar.
After much trial and error, I made a working method.
Mobile Phase: Acetonitrile/H20 75/25(v/v)
Column: Luna C18, 5u, 4.6 mm x 250 mm
Flow: 2.0 ml/min
Detector: UV at 254 nm
The retenion times for PBO is 6.228, Nylar is 6.858, I-Summithrn is 17.110, and II-Summithrn is 18.283 (all with good separation) The total run time is around 19 minutes.
My boss wants the total run time to be around 12 minutes. So I need to figure out how to get the retention time of summithrn down.
When I change the mobile phase to have more acetonitrile, it makes the total run time much lower, but PBO and Nylar peaks start to overlap each other. Same thing happens when I switch to a 150mm column.
Any ideas on what I should try next to lower the retention time would be greatly appreciated.
David
Flow: 2.0 ml/min
David