Running niacin on HPLC
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
3 posts
Page 1 of 1
Hi everyone! I was just wondering if anyone has ever had problems with their Niacin peaks on HPLC coming out as hills instead of nice sharp peaks? Or if not, any advice on how to correct this? Here are my specs: mobile phase is 10% methanol/90% water/1.7g tetrabutylammonium dihydrogen phosphate(per 100mL). Flow rate is 1.2mL/min and we are using a c18 column from Thermo. The method we are using is JAOAC vol 70, #4, 1987. Thanks! ~M
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
In general, "hills" instead of sharp peaks is the result of lots of diffusion. This can take place between the column and detector, if a piece of tubing is not properly installed (resulting in a gap) or it can be the result of "channeling" in a column (decay of stationary phase allowing gaps or channels to form in it, diffusion results).
Changes in peak shape can also be the result of "dirt" on the column (dirt being defined as undissolved particulates small enough to pass through a syringe filter or salt if the aqueous phase of your mobile phase has too much dissolved solids for the highest organic concentration of your gradient to maintain in a dissolved state, or crud in the tert... salt that was not filtered out).
Doing a lengthy low to high water/organic flush with the column reversed (sometimes this helps, but should not be done w/ columns whose inlet frits are sized very differently from outlet frits) may help restore desired peak shape.
Changes in peak shape can also be the result of "dirt" on the column (dirt being defined as undissolved particulates small enough to pass through a syringe filter or salt if the aqueous phase of your mobile phase has too much dissolved solids for the highest organic concentration of your gradient to maintain in a dissolved state, or crud in the tert... salt that was not filtered out).
Doing a lengthy low to high water/organic flush with the column reversed (sometimes this helps, but should not be done w/ columns whose inlet frits are sized very differently from outlet frits) may help restore desired peak shape.
Thanks,
DR

DR

-
- Posts: 5
- Joined: Tue Apr 23, 2013 8:26 pm
Thanks! Ill be sure to try it!
3 posts
Page 1 of 1
Who is online
In total there are 24 users online :: 2 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], John Guajardo and 22 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], John Guajardo and 22 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science