GC-MS automatic run
Posted: Sun Apr 14, 2013 7:53 pm
by sry
Hi everyone,
I have a problem with my GC-MS (GC: Agilent 6890, MS: Leco): after each injection (with Combi PAL autosampler) I must run the GC by pressing the "Start" key on the GC panel. How can I automate GC runnig? Is there any setting in the LECO softwere or in GC panel? (actually after each injection, GC reamins in "Pre Run" until manually pressing the "Start" key" ) .
Thanks a lot.
Re: GC-MS automatic run
Posted: Mon Apr 15, 2013 3:02 am
by Don_Hilton
There should be a start connection for the CombiPAL wired at the back of the GC. The LECO also needs a start signal - and I dont remember the details. While I have set up several of these, it's been a few years. It seems to me that the configuration for this is in the LECO manuals. They shoud be in electronic format on the instrument computer.
Was the instrument just set up by a vendor?
Posted: Mon Apr 15, 2013 12:16 pm
by sry
The instrument has not been used for 5 years and now I start to work with it. I don't know about setting up history. Everything works good except the auto-run.
Posted: Mon Apr 15, 2013 2:24 pm
by Yama001
Our PAL has an "Interface" connector that connects to the "Remote" jack on the back of the 6890 GC (its a split cable that also splits to the MS - its an Agilent MS so that part may be different for the LECO).
Posted: Mon Apr 15, 2013 11:48 pm
by walshbig
I exactly have the very same problem and I am working on it. Let me know your email and I will let you know if I found the solution.
Posted: Tue Apr 16, 2013 5:55 am
by walshbig
Problem is solved. Mine is working ok now. What version of Chromatof you are using? Let me know you email and I will guide you through required steps but need some info not to make any risky steps.
Posted: Sat Apr 20, 2013 10:24 am
by sry
My problem is also solved. But I have other problems too:
1- My GC-MS has a FID detector. Can I use it in Chromotof software or it must be run with Agilent Chemstation? If so, how?
2- For sensitivity check, what compound should I use? The Leco manual suggested HCB at 2 ppb, but I haven't it.
my e-mail: a.johnm@yahoo.com
Posted: Sat Apr 20, 2013 12:46 pm
by Don_Hilton
If you wish to use the FID along with the Pegasus, you should be able to take that signal into ChromaTOF.
HCB is readily available as a pre-made environmental standard. I typically order a couple of vials at a time. Typically you need to dilute the standard, so one vial will last you a l-o-n-g time. You need to keep the original solution tightly capped.
Posted: Mon Apr 22, 2013 7:59 am
by sry
Dear Walshbig
My problem also exist. I though my problem is solved but I have it again. I was wondering if you could help me.
GC: Agilent 6890
ChromaTOF Version:4.40
TOF version: Pegasus GC-HRT
Combi PAL: from Varian company without Head space and SPME tray (only for solution direct injection).
Posted: Tue Apr 23, 2013 2:21 am
by walshbig
I sent you an email. Check your email.
Posted: Tue Apr 23, 2013 8:58 am
by walshbig
There is a problem with your email. Please check and let me know if it is fixed. See bellow for the yahoo message:
Remote host said: 554 delivery error: dd This user doesn't have a yahoo.com account (a.johnm@yahoo.com) [0] - mta1343.mail.ne1.yahoo.com [BODY]
Posted: Thu Apr 25, 2013 7:52 pm
by sry
Dear Walshbig
You are in right. My email was defected. I'd really appreciate if you could mail me again (to my new ID: alcorn.mjohn@yahoo.com).
(problem: data aqusision is run only after manually pressing the "Satrt" button on the GC front pannel)
Thanks a lot.
Posted: Fri Apr 26, 2013 5:34 pm
by walshbig
Check you email