More recently we are working on a larger small molecule compound called Irganox 1010 with a molecular weight near 1178Da on a 6410QQQ, which is a larger non polar molecule (image attached). From review of the literature most people perform APCI + for better response, we have an ESI source, and I have also seen ESI + Ion trap performed to monitor the molecular ion. The mobile phase usually 20mM aqueous ammonium acetate and IPA as the organic ( I guess more non polar than ACN). The columns I have seen used are C18 so I prepared a 1,000 ng/mL standard and attempted a full scan injection without column to make sure we could detect the precursor ion, and we did which might indicate this compound ionizes to some extent. Then I plugged on our Agilent Zorbax C18 (2.1x100mm, 3.5micron) column with a guard column and attempted a very long ~30min scouting gradient, based on prior knowledge that this compound elutes nearly at 100% organic after quite a long time due to the large affinity towards stationary phase. I started the gradient at 85% organic and ramped to 100% over time. To make the story short, after the injection around 0.3mL/min, after a few minutes the system went overpressure, the lines popped off and a leak occurred, then it took me nearly 2 hours to rinse and restore the column to appropriate running conditions ( to lower the pressure). I believe this compound plugged my column because it was so highly retained. Since then I have been hesitant to reinject. Do you guys have any experience with this sort of situation, and how would you go about method development for this sort of compound? I have attached a figure of the compound .
Thanks so much